At 10:13 AM 10/16/2012, you wrote:
Quote: | Bob,
I ordered 3 of the devices. When comparing them at room temp they varied over a range of about 3 deg. Any info on what screw does what, or shall I just start twisting screws and see what happens ? I'm not as concerned with accuracy as I am with consistency across them for measuring temp differences. |
I've not explored the functionality of the two pots.
One will be offset (probably calibrated at the lower
temperature) and the other is scale factor (higher
calibration point).
You'll need a water-ice bath of distilled or RO
water and crushed ice. Also a boiling cup of
water . . .
Put dip into alternate baths, wait for stable
reading and then adjust pot for either 0 or
100 as appropriate. Each excursion/calibration
cycle between the two baths should display
smaller errors. If the errors are growing,
try reversing which pot is twisted for what
temperature.
Let us know what you discover!